RPRD | THE IMPORTANCE of FILTER INTEGRITY TEST to ENSURE STERILITY of RADIOP…

페이지 정보

작성자 : 서울대 핵의학과 작성일2017-02-02 조회3,898회

본문

Yong-Hyun Cho*, Jun-Hyung Park, Ki-Young Hwang, Hyung-woo Kim, Hong-Jae Lee, Hyun-Ju Kim

Pet center, Nuclear Medicine, Seoul National University Hospital, Seoul, Korea

 

 

Purpose

 

The radiopharmaceuticals are routinely injected to blood vessel for acquiring PET image. For this reason, It is imperative that they undergo strict quality control measures. Especially, Sterility test is more important than any other quality control procedures. Nevertheless, Due to radiation exposure and longer test time than half-lives of radionuclides, sterility test must be performed on a post hoc basis. According to the FDA guideline, It requires filter integrity test used in the processing of sterile solutions. Among several methods, we can decide to use bubble point test. This test will detect damaged filter membranes, ineffective seals, system leaks and so on.

 

 

Materials and Methods

 

We usually use vented GS-filters (Millipore co., USA) which are sterilizing-grade (0.22 um pore size) and are placed upper site on product vial. To prevent emitting radiation, a product vial is enveloped in the lead vial shield which has lead glass. During the synthesis of 18F-FDG, all of the solutions are passed through from ion exchange column, Sep-Paks, filter, to product vial by gas. Solutions wet the membrane in filter and then go into the product vial. By all synthesis steps have finished, we can observe the presence of the bubbles in the product vial.

 

 

Results

 

Since we have started this study, we have never found any bubbles in the product vial. Because the maximum pressure intensity of the filter which has set by manufacturer is up to 5 bars, but helium gas pressure is up to 1 bar in our module system. So, we can make 5 bars pressure using helium gas bombe and increase pressure up to 5 bars step by step. However, it does not happen to anything in vial.

 

 

Conclusion

 

The most widely used non-destructive filter integrity test is the bubble point test. It is a sensitive, visual technique. It must be routinely performed for radiopharmaceuticals sterility. In accordance with each test report, operators can decide whether it could dispense or not.

 

Key words: filter integrity test, bubble point test, non-destructive test, sterility, 0.22 um pore size.



서울특별시 종로구 대학로 101 (연건동 28)
진료예약안내 | 1588-5700
Copyrightⓒ 2017 Department of Nuclear Medicine Seoul National University Hospital. All Rights Reserved.
환자권리장전
서울대학교병원은 환자중심,인간존중,지식창조,사회봉사라는 경영이념을 통해 신체적/정신적 어려움으로 병원을 찾는 모든 환자의 권리를 존중하고, 최선의 진료를 제공하기 위하여 다음과 같이 선언한다.
존엄의 권리
환자는 존엄한 인간으로서 예우받을 권리가 있다.
평등의 권리
환자는 성별, 연령, 종교 또는 사회적 신분을 떠나 평등한 진료를 받을 권리가 있다.
설명을 들을 권리
환자는 의료진으로부터 질병의 진단,치료계획,결과,예후에 대한 설명을 들을 권리가 있다.
선택의 권리
환자는 치료,검사,수술,입원 등 본인에게 이루어지는 의료행위에 대한 설명을 듣고 시행여부를 선택할 권리가 있다.
개인신상 비밀을 보호받을 권리
환자는 진료내용,신체의 비밀 및 개인생활의 비밀을 보호받을 권리가 있다.