PET | Variation of measurements of standardized uptake value (SUV) using dif…
페이지 정보
작성자 : 서울대 핵의학과 작성일2017-02-02 조회3,096회관련링크
본문
1Tae Yeob Kim, 2Joong Hyun Kim, 1Jeom Jin Lim, 1Hong Jae Lee, 1Hyun Joo Kim, 1Jae Sung Lee,
1Dong Soo Lee
1Dept. of Nuclear Medicine, Seoul National Univ. Hospital
2Dept. of Nuclear Medicine, Seoul National U
Objective
Standardized uptake value (SUV) is a widely used semi-quantitative index in PET image analysis for the estimation of radiotracer accumulation in an interested region or volume. As mechanical characteristics of scanners are critical in the measurement of SUV, variation between different scanners should be considered in big institutes. In this study, SUVs of a phantom were measured using three different PET/CT scanners, and variation between them was evaluated.
Methods
The PET/CT scanners which were assessed in this study were GEMINI, GEMINI TF64 (Philips) and Biograph TruePoint 40 (Siemens Medical). A NEMA PET phantom (Data Spectrum Corp., USA) was used to measure SUVs. The phantom has weight of 6.8 kg and three hot inserts. Two sets of activity distributions for the background and inserts were tested; first, 3.7:3.7:7.4:11.1 MBq (1:1:2:3), and second, 1.85:7.4:9.25:11.1MBq (1:4:5:6), for background, insert 1, insert 2 and insert 3, respectively. Images were acquired for 2 bed positions with scan time of 2 min for each bed. SUVs for background and each insert were measured using each PET/CT scanner and user-provided software. The measurements of SUV were compared between scanners and with the theoretically calculated true values.
Results
For both the sets of activity distributions, all the scanners showed similar results. The variation between each scanner was statistically insignificant. The measurements of SUV were 91.2%, 85.9% and 87.2% of true SUVs, for GEMINI, GEMINI TF64, Biograph TruePoint 40, respectively.
Conclusion
For all the tested scanners, variation between the scanners was insignificant although the measured SUVs were slightly lower than the true values. Therefore, measurements of SUV using the tested scanners would be interchangeable in big institutes operating diverse PET/CT scanners. However, clinical validation is required as further studies