invitro | CA 19-9 검체량에 따른 농도의 변화
페이지 정보
작성자 : 서울대 핵의학과 작성일2017-02-04 조회4,063회관련링크
본문
서울대학교병원 핵의학과
장현영, 박희원, 유태민, 노경운, 김현주
Change of the Result Value by the Amount of Samples in CA 19-9
Hyun Yeong Chang, Hee Won Park, Tae Min You, Kyung Woon Noh, Hyun Joo Kim Department of Nuclear Medicine, Seoul National University Hospital, Seoul, Korea
Purpose: CA 19-9 need to examine a number of sample volume, and the postwar 200 U/ml concentration hook effect appears slight. Thus, the antibody-antigen reaction, and by reducing the amount of (sample volume), they can hook effect to minimize the impact of the sample volume and relevance know, I saw the hook effect. In addition, the current maximum of using the standard concentration of the reagent in 240 U/ml increase more than the standard concentration can be seen knows. Material and Method: 5 U/ml and under, make a few low concentration of serum pool from the high concentration of the sample hook effect together with a standard concentration of about 500 to meet the production. The reagents used in experiments are currently using SNUH NM experiment. Orignal method along with the experiment is to 25 ul sample volume (1 / 4), 50 ul (1 / 2), 100 ul (Orignal method) in the experiment. My greatest concentration of the reagent concentration of approximately two times the standard concentration of production. When was the last to make the first experiment, as measured by the standard concentration after that. The new inspection information through a standard solution modified by entering values in them. Results: 100 ul, and to apply the new standard concentration y = 1.3021x - 10.97, R2 = 0.9844. Overall, the results showed a similar orignal method. Because of the concentration in the value of more than 240 U/ml, but it is an overall value that can be made out of a similar value When I put the 50 ul y = 1.045x + 9.5861, R2 = 0.9428. Overall orignal method and the results of a similar value. 50 ul, and to apply the new standard oncentrationy=1.2006x+11.252, R2=0.9423. Showing a slightly lower value compared with orignal method. Because of the concentration in the value of more than 240 U/ml, but it is an overall value that can be made out of a similar value. When I put the man 25 ul y=0.6012x+24.755, R2=0.4033. Results showed that very small amounts of sample are insecure inside and showed a lower middle cpm orignal method and showed a lot of mismatched. Conclusions: 25 ul of the sample volume is not possible to use the instability had, when I put the 50 ul of the orignal method can be used to show a similar concentration. The new values are slightly lower concentration, The new values are slightly lower concentration, concentration, which are likely due to the lack of data has had a little gap between the sample showed 80 to 200 U/ml additional experiments seem to do. Apply a new 100 ul concentration values are applied to a large crowd is not even in sight. But this way the concentration of 100 to more 400 U/ml gather further experiments should possible adds. (Korean J Nucl Med Technol 2008;12(1):66-73)
Key Words : CA 19-9, Max dose, Full serum, Change a sample volume
서 론
CA 19-9 검사는 많은 검체량을 필요로 하며 200 U/ml 전 후의 농도에서 약간의 hook effect가 나타난다. 따라서 항체 와 반응하는 항원의 양(검체량)을 줄여가며 hook effect의 영향을 최소화 시킬 수 있는 검체량과 hook effect의 연관성을 알아보았다. 또한 현재 사용하고 있는 시약의 최대 표준액 농도인 240 U/ml 이상으로 표준액 농도를 더 높일 수 있는지 를 알아보았다.
대상 및 방법
몇 개의 풀 혈청을 만들어 5 U/ml 이하 저농도에서 고농도의 샘플을 모아서 hook effect 실험과 함께 표준액 농도를 약 500 U/ml 정도에 맞추어 제작한다. 실험은 현재 본원에서 사용하고 있는 키트를 이용하여 실험을 하였다. 실험은 원법과 더불어 샘플 양을 25 ul (1/4), 50 ul (1/2), 100 ul (원법)로 실험한다. 시약 내 최고농도의 약 2배 정도의 농도로 표준액을 제작한다. 처음 실험 때 가장 마지막에 만든 표준액 농도를 측정한 후에, 검사정보 수정을 통해서 새로운 표준액을 입력하여 값 을 낸다.
결 과
1. 100 ul와 새로운 표준액 농도를 적용
y = 1.3021x - 10.97, R2 = 0.9844 전반적으로 원법과 비슷한 결과를 보였다. 240 U/ml 이상의 농도에서 값이 나왔기 때문에 추세선 값이 안 좋아 보이지만 전반적으로 비슷한 값이 나왔다고 할 수 있다(Table 4, Fig. 4).
Fig. 1. Normal Method and 25 ul Method.
2. 50 ul만 넣었을 때 y = 1.045x + 9.5861, R2 = 0.9428 전반적으로 원법과 비슷한 결과 값이 나왔다(Table 2, Fig. 2).
2. 50 ul만 넣었을 때 y = 1.045x + 9.5861, R2 = 0.9428 전반적으로 원법과 비슷한 결과 값이 나왔다(Table 2, Fig. 2).
3. 50 ul와 새로운 표준액 농도를 적용
y = 1.2006x + 11.252, R2 = 0.9423 원법과 비교하여 약간 낮은 값을 보이고 있다. 240 U/ml 이상의 농도에서 값이 나왔기 때문에 추세선 값이 안 좋아 보이지만 전반적으로 비슷한 값이 나왔다고 할 수 있다(Table 3, Fig. 3).
4. 25 ul만 넣었을 때
y = 0.6012x + 24.755, R2 = 0.4033 매우 불안한 결과를 보였다 미량의 검체가 들어가다 보니 낮은 cpm을 보인 가운데 원법과 일치하지 않는 모습을 많이 보였다(자동화 장비에 의한 Carry Over로 보임) (Table 1, Fig. 1).
5. Hook effect 결과는 검체량을 줄였을 때
(1/2)가 원법보다 적은 hook effect을 보였다. 원법을 포함 한 다른 방법에서는 최대값 cpm이 높기에 1,000,000 U/ml 이상의 농도에서 원액이 hook effect을 보였다. 이는 항원 량을 적게 넣음으로 고농도의 검체에서 hook effect가 덜 일어났다.
Table 1. Normal Method and 25 ul Method
25 ul | 100 ul | |||
희석배수 | CPM | Dose | CPM | Dose |
STD1 | 675 | 0 | 648 | 0 |
STD2 | 856 | 15 | 1,861 | 15 |
STD3 | 1,318 | 30 | 3,976 | 30 |
STD4 | 2,606 | 60 | 7,900 | 60 |
STD5 | 5,261 | 120 | 14,984 | 120 |
Table 1. 계속
25 ul | 100 ul | |||
희석배수 | CPM | Dose | CPM | Dose |
STD6 | 7,531 | 240 | 19,926 | 240 |
Ctrl1 | 3,834 | 89 | 10,660 | 83 |
Ctrl2 | 1,716 | 40 | 5,677 | 43 |
1 | 441 | 1 | 551 | 1 |
2 | 500 | 1 | 1,054 | 7 |
3 | 770 | 7 | 1,176 | 8 |
4 | 789 | 9 | 1,372 | 11 |
5 | 1,333 | 30 | 2,186 | 18 |
6 | 2,994 | 69 | 2,756 | 22 |
7 | 5,501 | 131 | 2,959 | 23 |
8 | 6,443 | 178 | 5,493 | 42 |
9 | 3,439 | 79 | 5,285 | 40 |
10 | 1,808 | 42 | 6,481 | 49 |
11 | 470 | 1 | 6,250 | 48 |
12 | 477 | 1 | 10,945 | 86 |
13 | 2,596 | 59 | 15,369 | 128 |
14 | 3,245 | 75 | 14,448 | 115 |
15 | 5,636 | 137 | 17337 | 171 |
16 | 7,657 | 240 | 22,695 | 240 |
17 *1 | 10,506 | 23,119 | 22,695 | 520 |
18 *10 | 1,936 | 6,865 | ||
19 *100 | 771 | 1,378 | ||
20 *1 | 7,992 | 400 | 26,837 | 550 |
21 *10 | 1690 | 7,226 | ||
22 *100 | 895 | 5,951 | ||
23 *1 | 7,258 | 510 | 25,602 | 590 |
24 *10 | 2,203 | 7,819 | ||
25*100 | 1,182 | 1,524 |
Table 1. 계속
25 ul | 100 ul | |||
희석배수 | CPM | Dose | CPM | Dose |
26 *1 | 17,896 | 930 | 30,968 | 1,130 |
27 *10 | 4,041 | 14,156 | ||
28 *100 | 1,198 | 2,414 | ||
29 *1 | 28,930 | 3,600 | 37,917 | 4,400 |
31 *100 | 1,558 | 5,748 | ||
32 *1 | 34,239 | 13,300 | 39,496 | 13,800 |
33 | *10 | 35,283 | ||
34 *100 | 5,559 | 15,886 | ||
35 *1 | 36,385 | 43,000 | 32,900 | 46,000 |
36 *10 | 34,616 | 38,752 | ||
37 *100 | 11,395 | 26,292 | ||
38 *1,000 | 1,852 | 5,963 | ||
39 *10,000 | 752 | 1,234 | ||
40 *1 | 46,586 | 140,000 | 40,378 | 57,000 |
41 *10 | 35,340 | 40,734 | ||
42 *100 | 14,959 | 28,771 | ||
43 *1,000 | 5,706 | 7,546 | ||
44 *10,000 | 4,283 | 1,352 | ||
45 *1 | 44,781 | >2,400,000 | 37,406 | 112,000 |
46 *10 | 44,447 | 44,698 | ||
47 *100 | 37,055 | 34,532 | ||
48 *1,000 | 27162 | 14,010 | ||
49 *10,000 | 23467 | 2,859 | ||
50 *1 | 21,775 | >2,400,000 | 18,393 | 1,510,000 |
51 *10 | 45,797 | 29,211 | ||
52 *100 | 43483 | 43,912 | ||
53 *1,000 | 22,562 | 37,982 |
Fig. 2. Normal Method and 50 ul Method.
Table 2. Normal Method and 50 ul Method
50 ul | 100 ul | |||
희석배수 | CPM | Dose | CPM | Dose |
STD1 | 597 | 0 | 648 | 0 |
STD2 | 1,066 | 15 | 1,861 | 15 |
STD3 | 1,908 | 30 | 3,976 | 30 |
STD4 | 3,964 | 60 | 7,900 | 60 |
31 *100 | 1,558 | 5,748 | ||
STD5 | 7,387 | 120 | 14,984 | 120 |
STD6 | 11,514 | 240 | 19,926 | 240 |
Ctrl1 | 5,134 | 81 | 10,660 | 83 |
Ctrl2 | 2,702 | 42 | 5,677 | 43 |
1 | 433 | 1 | 551 | 1 |
2 | 596 | 1 | 1,054 | 7 |
3 | 596 | 1 | 1,176 | 8 |
4 | 675 | 3 | 1,372 | 11 |
5 | 909 | 11 | 2,186 | 18 |
6 | 1,371 | 21 | 2,756 | 22 |
7 | 1,673 | 26 | 2,959 | 23 |
8 | 2,947 | 46 | 5,493 | 42 |
9 | 2,506 | 40 | 5,285 | 40 |
10 | 2,555 | >40 | 6,481 | 49 |
11 | 2,147 | 34 | 6,250 | 48 |
12 | 3,553 | 55 | 10,945 | 86 |
Table 2. 계속
50 ul | 100 ul | |||
희석배수 | CPM | Dose | CPM | Dose |
STD1 | 597 | 0 | 648 | 0 |
13 | 1,066 | 15 | 1,861 | 15 |
STD3 | 4,612 | 30 | 3,976 | 30 |
STD4 | 3,964 | 71 | 15,369 | 128 |
14 | 6,305 | 101 | 14,448 | 115 |
15 | 8,017 | 137 | 17,337 | 171 |
16 | 11,713 | 240 | 22,695 | 240 |
17 *1 | 14,290 | 540 | 23,119 | 520 |
18 *10 | 3,511 | 6,865 | ||
19 *100 | 873 | 1,378 | ||
20 *1 | 16,103 | 450 | 26,837 | 550 |
21 *10 | 2,875 | 7,226 | ||
22 *100 | 987 | 5,951 | ||
23 *1 | 9,465 | 520 | 25,602 | 590 |
24 *10 | 3,354 | 7,819 | ||
25 *100 | 903 | 1,524 | ||
26 *1 | 21,460 | 1,100 | 30,968 | 1,130 |
27 *10 | 1,455 | 2,414 | ||
28 *100 | 2,555 | >40 | 6,481 | 49 |
29 *1 | 35,028 | 4,500 | 37,917 | 4,400 |
31 *100 | 2,864 | 5,748 | ||
32 *1 | 36,303 | 16,100 | 39,496 | 13,800 |
33 *10 | 29,214 | 35,283 | ||
34 *100 | 8,916 | 15,886 | ||
35 *1 | 32,068 | 48,000 | 32,900 | 46,000 |
36 *10 | 36,873 | 38,752 | ||
37 *100 | 16,718 | 26,292 | ||
38 *1,000 | 3,071 | 5,963 | ||
39 *10,000 | 936 | 1,234 | ||
40 *1 | 44,393 | 620,000 | 40,378 | 57,000 |
41 *10 | 39,292 | 40,734 | ||
42 *100 | 21,260 | 28,771 |
Table 2. 계속
50 ul | 100 ul | |||
희석배수 | CPM | Dose | CPM | Dose |
43 *1,000 | 4,064 | 7,546 | ||
44 *10,000 | 937 | 1,352 | ||
45 *1 | 40,736 | 132,000 | 37,406 | 112,000 |
46 *10 | 44,121 | 44,698 | ||
47 *100 | 32,869 | 34,532 | ||
48 *1,000 | 7,856 | 14,010 |
Table 2. 계속
50 ul | 100 ul | |||
희석배수 | CPM | Dose | CPM | Dose |
49 *10,000 | 1,287 | 2,859 | ||
50 *1 | 17,132 | 1,630,000 | 18,393 | 1,510,000 |
51 *10 | 37,968 | 29,211 | ||
52 *100 | 46,068 | 43,912 | ||
53 *1,000 | 33,008 | 37,982 | ||
54 *10,000 | 8,977 | 16,443 |
Fig. 3. Normal Method and 50 ul New Standard Method.
Table 3. Normal Method and 50 ul New Standard Method
50 ul | 100 ul | |||
희석배수 | CPM | Dose | CPM | Dose |
STD1 | 992 | 0 | 648 | 0 |
STD2 | 1,420 | 17 | 1,861 | 15 |
STD3 | 2,239 | 31 | 3,976 | 30 |
STD4 | 4,048 | 50 | 7,900 | 60 |
STD5 | 6,809 | 96 | 14,984 | 120 |
STD6 | 14,415 | 271 | 19,926 | 240 |
STD7 | 18,437 | 473 | ||
Ctrl1 | 5,134 | 68 | 10,660 | 83 |
Ctrl2 | 2,702 | 37 | 5,677 | 43 |
1 | 433 | 1 | 551 | 1 |
2 | 596 | 1 | 1,054 | 7 |
Table 3. 계속
50 ul | 100 ul | |||
희석배수 | CPM | Dose | CPM | Dose |
3 | 596 | 1 | 1,176 | 8 |
4 | 675 | 1 | 1,372 | 11 |
5 | 909 | 1 | 2,186 | 18 |
6 | 1,371 | 16 | 2,756 | 22 |
7 | 1,673 | 22 | 2,959 | 23 |
8 | 2,947 | 39 | 5,493 | 42 |
9 | 2,506 | 34 | 5,285 | 40 |
10 | 2,555 | 35 | 6,481 | 49 |
11 | 2,147 | 30 | 6250 | 48 |
12 | 3,553 | 46 | 10,945 | 86 |
13 | 4,612 | 59 | 15,369 | 128 |
Table 3. 계속
50 ul | 100 ul | |||
희석배수 | CPM | Dose | CPM | Dose |
14 | 6,305 | 88 | 14,448 | 115 |
15 | 8,017 | 121 | 17,337 | 171 |
16 | 11,713 | 205 | 22,695 | 240 |
17 *1 | 14,290 | 450 | 23,119 | 520 |
18 *10 | 3,511 | 6,865 | ||
20 *1 | 16,103 | 349 | 26,837 | 550 |
21 *10 | 2,875 | 7,226 | ||
22 *100 | 987 | 5,951 | ||
23 *1 | 9,465 | 153 | 25,602 | 590 |
24 *10 | 3,354 | 440 | 7,819 | |
25 *100 | 903 | 1,524 | ||
26 *1 | 21,460 | 960 | 30,968 | 1,130 |
27 *10 | 6,807 | 14,156 | ||
28 *100 | 1,455 | 2,414 | ||
29 *1 | 35,028 | 3,120 | 37,917 | 4,400 |
30 *10 | 15,340 | 24,232 | ||
31 *100 | 2,864 | 5,748 | ||
32 *1 | 36,303 | 14,100 | 39,496 | 13,800 |
33 *10 | 29,214 | 35,283 | ||
34 *100 | 8,916 | 15,886 |
Table 3. 계속
50 ul | 100 ul | |||
희석배수 | CPM | Dose | CPM | Dose |
35 *1 | 32,068 | 37,900 | 32,900 | 46,000 |
36 *10 | 36,873 | 38,752 | ||
37 *100 | 16,718 | 26,292 | ||
38 *1,000 | 3,071 | 5,963 | ||
39 *10,000 | 936 | 1,234 | ||
40 *1 | 44,393 | 50,000 | 40,378 | 57,000 |
41 *10 | 39,292 | 40,734 | ||
42 *100 | 21,260 | 28,771 | ||
43 *1000 | 4,064 | 7,546 | ||
44 *10,000 | 937 | 1,352 | ||
45 *1 | 40,736 | 118,000 | 37,406 | 112,000 |
46 *10 | 44,121 | 44,698 | ||
47 *100 | 32,869 | 34,532 | ||
48 *1,000 | 7,856 | 14,010 | ||
49 *10,000 | 1,287 | 2,859 | ||
50 *1 | 17,132 | 1,420,000 | 18,393 | 1,510,000 |
52 *100 | 46,068 | 43,912 | ||
53 *1,000 | 33,008 | 37,982 | ||
54 *10,000 | 8,977 | 16,443 |
Fig. 4. Normal Method and 100 ul New Standard Method.
Table 4. Normal Method and 100 ul New Standard Method
100 ul (2) | 100 ul | |||
희석배수 | CPM | Dose | CPM | Dose |
STD1 | 982 | 0 | 648 | 0 |
STD2 | 2,126 | 17 | 1,861 | 15 |
STD3 | 4,046 | 31 | 3,976 | 30 |
STD4 | 665 | 50 | 7,900 | 60 |
STD5 | 12,218 | 69 | 14,984 | 120 |
STD6 | 20,945 | 271 | 19,926 | 240 |
STD7 | 26,416 | 473 | ||
Ctrl1 | 10,660 | 83 | 10,660 | |
Ctrl2 | 5,677 | 43 | 5,677 | 43 |
1 | 551 | 1 | 551 | 1 |
2 | 1,054 | 3 | 1,054 | 7 |
3 | 1,176 | 6 | 1,176 | 8 |
4 | 1,372 | 9 | 1,372 | 11 |
5 | 2,186 | 18 | 2,186 | 18 |
6 | 2,756 | 22 | 2,756 | 22 |
7 | 2,959 | 24 | 2,959 | 23 |
8 | 5,493 | 42 | 5,493 | 42 |
9 | 5,285 | 41 | 5,285 | 40 |
10 | 6,481 | 49 | 6,481 | 49 |
11 | 6,250 | 47 | 6,250 | 48 |
12 | 10,945 | 86 | 10,945 | 86 |
13 | 15,369 | 150 | 15,369 | 128 |
14 | 14,448 | 133 | 14,448 | 115 |
15 | 17,337 | 189 | 17,337 | 171 |
16 | 22,695 | 329 | 22,695 | 240 |
17 *1 | 23,119 | 520 | 23,119 | 520 |
18 *10 | 6,865 | 6,865 | ||
20 *1 | 26,837 | 550 | 26,837 | 550 |
21 *10 | 7,226 | 7,226 |
Table 4. 계속
100 ul (2) | 100 ul | |||
희석배수 | CPM | Dose | CPM | Dose |
22 *100 | 5,951 | 5,951 | ||
23 *1 | 25,602 | 439 | 25,602 | 590 |
24 *10 | 7,819 | 600 | 7,819 | |
25 *100 | 1,524 | 1,524 | ||
26 *1 | 30,968 | 1,280 | 30,968 | 1,130 |
27 *10 | 14,156 | 14,156 | ||
28 *100 | 2,414 | 2,414 | ||
29 *1 | 37,917 | 3,850 | 37,917 | 4,400 |
30 *10 | 24,232 | 24,232 | ||
31 *100 | 5,748 | 5,748 | ||
32 *1 | 39,496 | 16,000 | 39,496 | 13,800 |
33 *10 | 35,283 | 35,283 | ||
34 *100 | 15,886 | 15,886 | ||
35 *1 | 32,900 | 46,800 | 32,900 | 46,000 |
36 *10 | 38,752 | 38,752 | ||
37 *100 | 26,292 | 26,292 | ||
38 *1000 | 5,963 | 5,963 | ||
39 *10000 | 1,234 | 1,234 | ||
40 *1 | 40,378 | 58,000 | 40,378 | 57,000 |
41 *10 | 40,734 | 40,734 | ||
42 *100 | 28,771 | 28,771 | ||
43 *1000 | 7,546 | 7,546 | ||
44 *10000 | 1,352 | 1,352 | ||
45 *1 | 37,406 | 125,000 | 37,406 | 112,000 |
46 *10 | 44,698 | 44,698 | ||
47 *100 | 34,532 | 34,532 | ||
48 *1000 | 14,010 | 14,010 | ||
49 *10000 | 2,859 | 2,859 | ||
50 *1 | 18,393 | 1,710,000 | 18,393 | 1,510,000 |
Table 4. 계속
100 ul (2) | 100 ul | |||
희석배수 | CPM | Dose | CPM | Dose |
52 *100 | 43,912 | 43,912 | ||
53 *1000 | 37,982 | 37,982 | ||
54 *10000 | 16,443 | 16,443 |
결 론
검체량을 25 ul은 불안정하여 사용이 불가능하여 보였고, 50 ul을 넣었을 때는 원법과 비슷한 농도를 보이기에 사용이 가능 할 듯하다. 새로운 농도를 적용한 값은 약간 낮은 농도를 보이기 때문에 현재의 데이터로는 약간의 부족함이 보여 80~200값 사이의 샘플을 추가로 실험을 해야 할 듯하다. 100 ul에 새로운 농도를 적용한 값은 적용을 해도 큰 무리가 없어 보인다. 하지만 이 방법도 100~400 uU/ml까지의 농도를 더 모아서 추가 실험을 더 해봐야 할 듯하다.
고 찰
1. 본원에서 사용하고 있는 CA 19-9 시약은 특정한 값에서 hook effect를 보이는 경우가 종종 일어난다.
2. 결과적으로 항원량을 줄임으로 원법보다 더 낳은 결과 값을 산출할 수 있었고, 실험실 내에서 검체량 부족으로 인한 실험의 오류를 최소화할 수 있다고 본다.
3. 실전에 투입 전에 더 많은 실험을 통해서 조만간 본원에 이 방법을 적용할 계획이다.
참고문헌
1. 임상 핵의학검사 기술학. 고려의학. 1993;21:329~352.
2. Magnani J, Brocklaus M, Smith D, Ginsburg V, Blaszcyk M, Mitchell D, Steplewski Z, and Koprowski H. A monoclonal anitibody defines antigen of colon Calcinoma. Science. 1981;212:55.
3. Amide M, et al. CA 19-9 Producing testicular tumor: a case report, hinyokika kiyo, 2000;46(1):45-47.
4. Nemann V, et al. CA19-9 : a predictor of response in pancreatic cancer treated with gemcitabine and cisplatin. Anticancer Res. 1999;19(4a):2433-2435.
5. Steplewski Z, and Koprowski H. Federation Proceedings 1982; 41:726.
6. Delvilano B, Brennan S, Brock P, Bucher C, Liu V, McClure M, Rake B, Space S, Westrick B, Schoemakler H, Zurawski V. Radioimmnometric Assay for a Monoclonal Antibody-Defineed Tumor Maker CA19-9. Clin Chem. 1983;29:549.