Gamma | Study for automatic exposure control technique (AEC) in SPECT/CT for r…

페이지 정보

작성자 : 서울대 핵의학과 작성일2017-02-01 조회3,506회

본문

Seok Hwan Yoon, Seong wook Cho, Jin Eui Kim

Department of Nuclear Medicide, Seoul National University Hospital, Seoul, Korea,

Purpose : Auto exposure control (AEC) in SPECT/CT automatically controls the exposure dose (mA)

according to patient's shape and size. The aim of this study was to evaluate the effect of AEC in SPECT/CT

on exposure dose reduction and image quality. Methods: The model of SPECT/CT used in this study was

Discovery 670 of GE Co.; Smart mA for AEC; and 99mTc as a radioisotope. To compare SPECT and CT

images by CT exposure dose variation, we used a standard technique set at 80/100/120/140kVp,

10/30/50/100/150/200/250mA, and AEC at 80/100/120/140kVp, 10~250mA. To evaluate resolution and contrast

of SPECT images, triple line phantom and flangeless Esser PET phantom were used. For CT images, noise

and uniformity were checked by anthropomorphic chest phantom. For dose evaluation to find DLP value,

anthropomorphic chest phantom was used and the CT protocol of torso was applied by standard technique

(120kVp, 100mA) and AEC (120kVp, 10~250mA).

Results : When standard and AEC were applied, the resolutions at SPECT images with Attenuation

Correction(A.C) were the same as FWHM by Center 3.65㎜ , Left 3.48㎜ , Right 3.61㎜ . Contrasts of

standard and AEC showed no significant difference: standard 53.5/29.8/22.5/15.8/6.0, AEC

53.5/29.6/22.4/15.7/6.1 In CT images, noise values at standard and AEC were 15.4 and 18.5 respectively. The

application of AEC increases noise but the value of coefficient variation were 33.8, 24.9 respectively,

obtaining uniform noise image. The values of DLP at standard and AEC were 426.78 and 352.09 each, which

shows that the application of AEC decreases exposure dose more than standard by approximately 18%.

Conclusions : The results of our study show that there was no difference of A.C in SPECT images based on

the CT exposure dose variation at SPECT/CT images. It was found that the increased CT exposure dose leads

to the improvement of CT image quality but also increases the exposure dose. Thus, the use of AEC in

SPECT/CT contributes to obtaining equal A.C SPECT images, and uniform noise in CT images while

reducing exposure dose.

Key Words : SPECT/CT, AEC(Automatic Exposure Control), Attenuation correction


서울특별시 종로구 대학로 101 (연건동 28)
진료예약안내 | 1588-5700
Copyrightⓒ 2017 Department of Nuclear Medicine Seoul National University Hospital. All Rights Reserved.
환자권리장전
서울대학교병원은 환자중심,인간존중,지식창조,사회봉사라는 경영이념을 통해 신체적/정신적 어려움으로 병원을 찾는 모든 환자의 권리를 존중하고, 최선의 진료를 제공하기 위하여 다음과 같이 선언한다.
존엄의 권리
환자는 존엄한 인간으로서 예우받을 권리가 있다.
평등의 권리
환자는 성별, 연령, 종교 또는 사회적 신분을 떠나 평등한 진료를 받을 권리가 있다.
설명을 들을 권리
환자는 의료진으로부터 질병의 진단,치료계획,결과,예후에 대한 설명을 들을 권리가 있다.
선택의 권리
환자는 치료,검사,수술,입원 등 본인에게 이루어지는 의료행위에 대한 설명을 듣고 시행여부를 선택할 권리가 있다.
개인신상 비밀을 보호받을 권리
환자는 진료내용,신체의 비밀 및 개인생활의 비밀을 보호받을 권리가 있다.