Gamma | Gamma Camera에 있어 측면 선란선의 영향에 대한 평가

페이지 정보

작성자 : NMSNUH 작성일2021-05-25 조회634회

본문

 

The Evaluation of Lateral Scatter Ray of Gamma Camera

 

Jae-Il Kim, Eun-Byeol Lee, Seong-Wook Cho, Kyeong-Woon Noh and Keon-Wook Kang Department of Nuclear Medicine, Seoul National University Hospital, Seoul, Korea

 

Purpose : Generally, a collimator that installed in front of detector set a direction of gamma ray and remove a scatter ray. By the way, a lateral or oblique scatter ray is detected into crystal through collimator. At this study, we will evaluate a mount of count and spectrums of lateral scatter ray.

 

Materials and Methods : We used the SKY LITE (philips, netherlands) as a gamma camera, and 99mTc, 1.11 GBq point source as a phantom. we put this point source at backside 50 of detector. After acquiring this for 1 min, we turned a detector next 10 degrees. Likely this, we acquired images at every 10 degrees from 0° to 360°, analyzed images and spectrums. In case of patient study, we choose a 3 phase bone scan patient who had a hand disease, because scatter rays from body would detect on crystal. After acquiring blood flow and blood pool images, we analyzed images and spectrums. Additional, we put a lead gown on patient’s hand, body. And then we compared and evaluated 3 type blood pool images (non lead gown, lead gown on a hand and on body).

 

Results : In case of phantom study, scatter ray counts at backside (270°-90°) are same with a background count. By the way, counts of scatter ray of oblique side (0°-50°, 220°-270°) are 100-600 cps, furthermore, counts at frontside are over 4 Mcps. In case of patient study, a counts of hand blood pool scan are 1510 cps. But counts of hand with lead gown on hands and on body are each 1554 cps, 1299 cps.

 

Conclusion : Therefore, even though there is a collimator in front of detector, lateral scatter rays detect on crystal and affect to images and spectrums. Especially, if there is a high activity source at outside of detector when we examine low activity organs like hands or foot, we have to shield and remove the source at outside for a good image.

 

Key Words Gamma Camera, Scatter Ray, Collimator, Spectrum


서울특별시 종로구 대학로 101 (연건동 28)
진료예약안내 | 1588-5700
Copyrightⓒ 2017 Department of Nuclear Medicine Seoul National University Hospital. All Rights Reserved.
환자권리장전
서울대학교병원은 환자중심,인간존중,지식창조,사회봉사라는 경영이념을 통해 신체적/정신적 어려움으로 병원을 찾는 모든 환자의 권리를 존중하고, 최선의 진료를 제공하기 위하여 다음과 같이 선언한다.
존엄의 권리
환자는 존엄한 인간으로서 예우받을 권리가 있다.
평등의 권리
환자는 성별, 연령, 종교 또는 사회적 신분을 떠나 평등한 진료를 받을 권리가 있다.
설명을 들을 권리
환자는 의료진으로부터 질병의 진단,치료계획,결과,예후에 대한 설명을 들을 권리가 있다.
선택의 권리
환자는 치료,검사,수술,입원 등 본인에게 이루어지는 의료행위에 대한 설명을 듣고 시행여부를 선택할 권리가 있다.
개인신상 비밀을 보호받을 권리
환자는 진료내용,신체의 비밀 및 개인생활의 비밀을 보호받을 권리가 있다.